When the movable arm of a Michelson interferometer in vacuum (n = 1) is moved by 325 μm, the number of fringe crossings is 1000. The wavelength of the laser used in nanometers is ____.

Correct Answer

650-650

Solution & Step-by-step Explanation

Concept:-

Applications of Michelson interferometer

(i) Find the wavelength of the monochromatic light

nλ = 2 (Δx)

Δx = x₂ - x₁

where, n = no. of fringes

λ = wavelength of monochromatic light

Δx = micrometer screw movement

x₂ = final position, x₁ = initial position

(ii) Calculation of thickness and the refractive index (μ) of any transparent sheet.

image

nλ = 2t [μ - 1]

t = thickness of sheet

μ = Refractive index of steel material.

(iii) Resolution of spectral lines.



(iv) calculation of thickness of sheet width by young's method

image

nλ = t (μ - 1)

thickness 't' =

Calculation:-

Given, μ = 1

Δx = 325 μm

n = 1000

λ = ?

By the first application of michelson interferometer

nλ = 2 + (μ - 1) = 2Δx

nλ = 2Δx

1000 × λ = 2 × 325 × 10⁻⁶

λ = 650 × 10⁻⁹

λ = 650 nm.

Practice this question

Try it yourself before checking the explanation above.

When the movable arm of a Michelson interferometer in vacuum (n = 1) is moved by 325 μm, the number of fringe crossings is 1000. The wavelength of the laser used in nanometers is ____.

Share This Question

Related Questions

mediumMCQ

A 2n × 2n matrix A=[a_ij] has its elements as a_ij = β & if (i+j) is odd, \\ -β & if (i+j) is even, where n is any integer greater than 2 and β is any non-zero real number. The rank of A is

mediumMCQ

The solution of the differential equation dy/dx = 9x/y represents

mediumMCQ

The working of the hand-held metal detector most widely used by security personnel for human frisking is based on the principle of

mediumMCQ

The primary coil of a linear variable differential transformer (LVDT) is supplied with AC voltage as shown in the figure. The secondary coils are connected in series opposition and the output is measured using a true RMS voltmeter. The displacement x of the core is indicated in mm on a linear scale. At the null position x = 0 , the voltmeter reads 0 V. If the voltmeter reads 0.2 V for a displacement of x = +2 mm, then for a displacement of x =-3 mm, the voltmeter reading, in V, is ![image](https://cdn.prepp.in/uploads/production/images/content/2025_11_17/image_7cdc73f01763403640224.png)

mediumMCQ

In the force transducer shown in Figure (a), four identical strain gauges S1, S2, S3, and S4 are mounted on a cantilever at equal distance from its base. S1 and S2 are mounted on the top surface and S3 and S4 are mounted on the bottom surface, as shown in the Figure (a). These strain gauges are to be connected to form a Wheatstone bridge consisting of four arms A, B, C, and D, as shown in the Figure (b). From the following options, the correct order to maximize the measurement sensitivity is ![image](https://cdn.prepp.in/uploads/production/images/content/2025_11_17/image_b4a3a5441763403707068.png)

mediumMCQ

Let a continuous-time signal be x(t) = e^j9t+e^j5t , where j = √(-1) and t is in seconds. The fundamental period of magnitude of x(t) , in seconds, is

Ready for a Full Test?

Practice with timed mock tests and track your performance across Instrumentation Engineering.

Discussion